Viscosity
Fluids exhibit a definite resistance to change of form. This resistance to change represents a type of internal friction and is called viscosity. In practice the viscosity of a fluid usually manifests itself as a resistance to flow, both in liquids and gases but at a much lower level in gases. Viscosity is classically defined as the tangential force per unit area necessary to maintain unit relative velocity between two parallel plates in a liquid unit distance apart. The unit of viscosity is called the poise. Another form of viscosity measurement is the kinetic viscosity which is the ratio of the viscosity in poises to the fluid density in gram per ml. Kinetic viscosity is measured in stokes.
Author: RPW Scott
Book:The Thermodynamics of Chromatography
Section:Thermodynamics Other-Methods Chiral-Separations Optimum-Velocity
solvents, canbe calculated
using an equation developed by Moore et al. [20], viz,
(52)
where (y1)
is the mole fraction of solvent (1),
(y2)
is the mole fraction of solvent (2),
(h1)
is the viscosity of solvent (1),
and (h1)
is the viscosity of solvent 2.
By employing
the viscosity values calculated for a range of temperatures and solvent
mixtures in equation (50) the effect of temperature and solvent composition on
the optimum mobile phase velocity could be calculated. A practical initial
value for () was assumed to be 2.5 x 10-5 cm2s.
This value was taken as the probable estimate from publishd data for similar
compounds (21,22) for the diffusivity of
Thermodynamics Other-Methods Chiral-Separations Optimum-Velocity
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Van-Deemter-Equation
allowing a longer column to be used. The permeability increases as
the square of the particle diameter but the variance per unit length only
increases linearly with the particle diameter. Thus, doubling the particle
diameter will allow a column four times the length to be used but the number of
plates per unit length will be halved. Consequently, the column
efficiency will be increased by a factor of two. It is also seen that the
higher efficiencies will be obtained with mobile phases of low viscosity and
for solutes of low diffusivity. Solvent viscosity and solute diffusivity tend
to be inversely proportional to each other and so the sensitivity of the
maximum obtainable efficiency to either solvent viscosity or solute diffusivity
will generally not be large. The approximate length of a column that will
provide the maximum column efficiency when operated at optimum velocity is
given by, l = nHmin
Dispersion Van-Deemter-Equation
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Golay-Equation
nbsp;It is
interesting to note from equation (60) that when a capillary column is run at
its optimum velocity, the maximum efficiency attainable is directly
proportional to the inlet pressure and the square of the radius and inversely
proportional to the solvent viscosity and the diffusivity of the solute
in the mobile phase. This means that the maximum efficiency attainable from a
capillary column increases with the column
radius. Consequently, very high
efficiencies will be obtained from relatively large diameter, but
very long, columns.
Employing
equation (60) it is possible to calculate the range of efficiencies attainable
under operating conditions commonly used for packed columns. The conditions
assumed are as follows,
 
Dispersion Golay-Equation
Author: RPW Scott
Book:The Thermodynamics of Chromatography
Section:Thermodynamics Other-Methods Chiral-Separations Optimum-Velocity
nbsp;
(51)
where
is the Diffusivity of the solute at temperature (Tx) and solvent composition (cx)
is the Diffusivity of the solute at the reference temperature (To) and solvent composition (co)
is the viscosity of the solvent at temperature (Tx) and solvent composition (cx)
and
is the viscosity of the solvent at the reference temperature (To) and solvent composition (co
Thermodynamics Other-Methods Chiral-Separations Optimum-Velocity
Author: RPW Scott
Book:Gas Chromatography
Section:YES Gas-Supplies Flow-Controllers
Flow Controllers
A constant
pressure applied to a column does not ensure a constant flow of mobile phase
though the chromatographic system, particularly if the column is being
temperature programmed. Raising the temperature of a gas causes the viscosity
to increase, and at a constant inlet pressure, the flow rate will fall. The
reduction in flow rate will be related to the temperature program limits and to
a certain extent on the temperature gradient. To obviate the flow rate change,
mass controllers are used which ensure a constant mass of mobile passes through the column in unit time irrespective of the system
temperature. A diagram of a mass flow controller is shown in figure 3.
Courtesy of Porter
YES Gas-Supplies Flow-Controllers
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Van-Deemter-Equation
nbsp;
Thus,
(44)
It is seen
that the column length varies inversely as the product of the solute
diffusivity in the mobile phase and the mobile phase viscosity in much the same
way as the column efficiency does when operating at the optimum velocity. As
would be expected the column length is directly proportional to the inlet
pressure but, less obviously is also proportional to the cube of the particle
diameter.
The analysis
time for a solute mixture in which the last peak is eluted at a capacity ratio
of k'F is given by,
 
Dispersion Van-Deemter-Equation