Thermal Conductivity Detectors
Thermal conductivity detectors (or hot wire detectors as they are sometimes known) do not respond exclusively to changes in thermal conductivity of a gas as was first thought. The basic sensor of the thermal conductivity detector was originally developed for measuring the concentration of carbon dioxide in flue gasses. The concept was then borrowed for use as a detector in gas chromatography. The sensor consists of a filament of platinum or tungsten wire situated in the eluent gas and heated by an electrical current. In the presence of the carrier gas alone the wire comes into thermal equilibrium at a given temperature. If placed in the arms of a Wheatsone bridge in which another arm contains a similar sensor situated in a reference stream of the same gas the bridge can be electrically balanced. When solute vapor is eluted from the column, the thermal conductivity and the specific heat of the gas surrounding the heated wire, changes. This change in thermal properties of the system results in a change in heat loss from the wire, a consequent change in wire temperature and wire resistance and the bridge becomes out of balance. The out-of-balance signal is amplified and fed to a recorder. The response of the sensor to change in specific heat of the surrounding gas is at least as great as that due to the change in thermal conductivity. The thermal conductivity detector has a sensitivity of about 10-6 to10-7 g/ml, and a linear dynamic range of about three orders of magnitude.
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Early-GC-Detectors Flame-Thermocouple
the other in a pure
reference stream of gas. In the presence of a solute, both the thermal
conductivity and the heat capacity of the gas change changing the heat loss
and, thus, the temperature of the filament and, consequently, its resistance
The bridge is unbalanced and the out-of-balance signal is passed to a suitable
monitoring device. This detector is relatively insensitive but responds to all
solutes that differ in heat capacity and thermal conductivity from those of the
carrier gas. This detector was used extensively in the early days of GC for the
analysis of hydrocarbon gases. There was much discourse and dissent with
regards to the exact mechanism of detection involved in the katharometer and
even today it is considered to respond to a number of different physical
properties of the eluent gas with no one property playing a major role.
The Flame Thermocouple Detector
The "flame thermocouple detector"
was the next detector to be reported which was developed
GC-Detectors Early-GC-Detectors Flame-Thermocouple
Author: RPW Scott
Book:Gas Chromatography
Section:YES Detectors Katherometer
nbsp;
The Katherometer Detector
The
katherometer detector (sometimes spelt catherometer and often referred to as
the thermal conductivity detector or hot wire detector) is
relatively insensitive but has survived largely as a result of its catholic
response and, in particular, its response to the permanent gases. Consequently,
it is often the detector of choice for gas analysis and environmental testing.
Its frequent use in these special types of application, somewhat surprisingly,
has made it the fourth most commonly used GC detector. A filament carrying a
current is situated in the column eluent and, under equilibrium
YES Detectors Katherometer
Author: RPW Scott
Book:Preparative Chromatography
Section:Preparative Apparatus Detectors
chromatography
detectors can have large sensor volumes and, as the detector is required only
to monitor the separation, they need not have a linear response. They do need
to tolerate high flow rates and thus, must have low flow impedance. Analytical
detectors can be used for preparative purposes but a portion is usually split
from the column eluent, diluted with more mobile phase and then passed through
the detector. In practice this becomes a rather clumsy procedure.
The most
commonly used detector in preparative GC is the thermal conductivity detector
(hot wire detector). Even this detector, however, is often too sensitive and
has too high a flow impedance. Under such circumstances, the procedure
mentioned above must be employed. The eluent from the preparative column is
split and a small portion diverted through the detector (sometimes with further
dilution with carrier gas to reduce sensitivity).
In LC, the
refractive index detector is probably the most useful of the
Preparative Apparatus Detectors
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors General-Properties
of GC Detectors
The GC detector is designed to respond to
very small quantities of vapor contained in a permanent gas. Because the
physical and chemical properties of permanent gases differ widely from those of
a vapor, a very wide range of detection methods can be employed including the
measurement of standard physical properties such as thermal conductivity and
light adsorption to more specific properties such as ionization potentials and
heats of combustion.
The response of a GC detector can be general
or specific but a detector with a catholic response is generally more useful in
routine analyses. Aspecificdetector(e.g.,the nitrogen-phosphorus detector (NPD)) can be extremely useful for
selectively monitoring compounds such as herbicides and pesticides, when the
compounds are not eluted discretely but mixed with a number of other
contaminating compounds.
GC detectors should be insensitive to changes
in flow rate but, unfortunately, few detectors have this
GC-Detectors General-Properties
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Katharometer
filament assumes a constant temperature. The heat lost from
the filament will depend on both the thermal conductivity of the gas and its
specific heat. Both these parameters will change in the presence of a different
gas or solute vapor and as a result the temperature of the filament changes,
causing a change in potential across the filament. This potential change is
amplified and either fed to a suitablerecorder
or passed to an appropriate data acquisition system.As the detector filament is in thermal equilibrium with its
surroundings and the device actually responds to the heat lost from the
filament, the detector is extremely flow
and pressure sensitive. Consequently, all katharometer detectors must be carefully thermostatted and
must be fitted with reference cells to help compensate for changes in pressure
or flow rate.
Figure 13. The Off-Line
Katharometer Sensor
GC-Detectors Katharometer