Low Dispersion
Dispersion, in this chromatography context, refers to the capacity of a column or some part of the mobile phase conduit system to cause ‘peak dispersion’ or peak spreading. Lowdispersion indicates the contribution to peak spreading is small. It is a term usually given to describe parts of the mobile phase system other than the column. For example low dispersion unions are designed to ensure that dispersion that occurs as a solute band passes through them is strictly constrained. Low dispersion tubing is made to limit dispersion that can take place in connecting tubes between column and detector (an example of which is Serpentine tubing). There are also low dispersion sample valves and low dispersion detector sensor cells all of which are carefully designed to limit any peak spreading that may take place in them. It is extremely important to carefully restrict any peak dispersion that takes place outside the column as such dispersion can impair (and in some cases destroy) the separation that is achieved in the column.
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Unions-and-Frits
still persists. It would
appear that the low dispersion serpentine tubing is the most satisfactory
alternative to straight tube.
Dispersion in Unions and Stainless Steel Frits
Depending on
their design, unions can also be a significant source of extra column
dispersion. Instrument manufacturers have been aware of the problem of union
dispersion and, as a consequence, have designed low dead volume unions which
are now generally available. Actual data reporting the extent of the dispersion
that takes place in such unions does not, however, appear to be readily
available. Scott and Simpson also
measured the relative dispersion that occurred in normal, low dead volume
unions and drilled-out unions. Drilled-out unions allow the ends of the
connecting tubes to butt against one another, or against the frit of a
microbore column and thus reduce the union volume dead volume to virtually
zero. The design of low dead volume and drilled out unions are depicted in
figure 15
EC-Dispersion Unions-and-Frits
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Dispersion Low-Dispersion-Tubing
in only one commercial LC detector. It should be pointed out that any
conduit system that has low dispersion will also provide very fast heat
transfer rates. Serpentine tubing has been also used in commercial column ovens
to heat the mobile phase rapidly to the column oven temperature before it
enters the column. The serpentine tubing allows effective heat exchange with a
minimum of heat exchanger volume to distort the concentration profile of the
solvent gradient.
The different
forms of dispersion profiles that are obtained from various types of connecting
tubes used in LC are shown in figure 4.
Figure 4 Dispersion Profiles
from Different Types of Tube
These
dispersion curves were obtained using a low dispersion UV detector (cell
volume, 1.4 ml) and a sample valve with a 1 ml internal loop. All tubes were of the same
length and carried the same mobile phase at a flow rate of 2 ml/min. employed.
The peaks were recorded on a high speed recorder. The peak from
HPLC-Detectors Dispersion Low-Dispersion-Tubing
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Small-Bore-Columns
dispersion should be allotted to the sample volume, as a
large sample volume may be necessary to handle a particular sample type for
successful analysis. In any event, the chromatograph should always be designed
so that the dispersion from other parts of the system is kept to the absolute
minimum. Dispersion in sample valves can be minimized by mechanical design
(internal loop valves tend to provide the minimum dispersion). Dispersion from
unions can be minimized by using drilled-out unions or low dead volume unions.
Stainless
steel frits provide very little dispersion and can be employed without great
concern for their contribution to the overall dispersion of the system.
Connecting tubes are one of the major sources of extra-column dispersion and
should be kept as short as possible and the radius reduced to a minimum
commensurate with the tube not becoming blocked. Tubes 0.005 in. in diameter
are recommended as the minimum diameter that is unlikely to become blocked. In
EC-Dispersion Small-Bore-Columns
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Dispersion Low-Dispersion-Tubing
Low Dispersion Tubing
In order to
avoid dispersion in mobile phase conduits a number of attempts to design low
dispersion tubing has been reported. The first attempt was by Halasz et al.
(8), who crimped and bent the tube into different shapes to interrupt the
Newtonian flow and introduce radial flow within the tube. His devices had
limited success and the tubes had a tendency to block very easily.
In 1978
Tijssen (9), developed a theory to describe the radial flow that was induced
into coiled tubes by the continual change in direction of the fluid as it
flowed round the spirals (his theory will be
HPLC-Detectors Dispersion Low-Dispersion-Tubing
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion HPLC-Sample-Valves
0.030 in
0.667
0.010 in
0.338
Flow rate 20 ml/min.
The data was
obtained using a specially designed low dispersion chromatographic system and
an electrical conductivity detector with a sensor volume of only 0.08 ml.
Despite its
short length, it is clear that the internal bore of the valve connection can
still cause significant dispersion. In addition, it would appear that if the
diameter of the connection was further reduced the dispersion would be even
less. However, apertures less than 0.005 in. I.D. can readily become blocked.
The absolute minimum internal diameter that is recommended is 0.003 in. and if
apertures of such dimension are present in the chromatographic system all
samples and the mobile phase should be filtered before use. From a theoretical
point of view, it should be possible to introduce secondary flow into the
conduit, increase the solute diffusivity, and consequently, reduce dispersion.
The
EC-Dispersion HPLC-Sample-Valves
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Low-Dispersion-Connecting-Tubes
nbsp;
Low Dispersion Connecting Tubes
The ideal
solution to conduit dispersion, where the sample valve and the detector sensor
cell are coupled directly to the column, is, in practice, mostly impossible.
Consequently, a conduit system that provides little or no dispersion would be
extremely useful. In order to reduce dispersion due to Newtonian flow through
an open tube, the parabolic velocity profile of the fluid must be disrupted to
introduce rapid radial mixing. The parabolic velocity profile can be disturbed,
and secondary flow introduced, into the tube, by deforming its regular
geometry.
Dispersion
that occurs in geometrically deformed tubes (squeezed, twisted and coiled) has
been studied by Halasz (4, 5 and 6), and the effect of radial convection
(secondary flow) on the dispersion introduced in
EC-Dispersion Low-Dispersion-Connecting-Tubes