LC Detectors
There are a number of LC detectors available but only a relative few are in general analytical use. There are two basic types of LC detectors, ‘bulk property detectors’ and ‘solute property detectors’. The first measures some overall physical property of the column eluent (e.g. the refractive index detector responds to changes in the refractive index of the column eluent). The second respond to some unique property of the solute (e.g. the UV detector that responds to changes in transmitted light through the sample due to UV absorption). In general the ‘bulk property’ detectors are les sensitive than the ‘solute property’ detectors. LC detectors can exhibit a range of sensitivities from 10-6 g per ml (e.g. the refractive index detector) to about 10-10 g per ml (e.g. the electrochemical detector). All LC detectors must have a linear, or close to linear, response over a reasonable concentration range (e.g. three orders of magnitude) for effective use in quantitative analysis. LC detectors must also have relatively small sensor volumes (e.g. 5 micro-liters or less) to prevent peak dispersion and, thus, loss of resolution. The three most commonly used LC detectors are probably the UV absorption detector, the fluorescence detector and the electrical conductivity detector.
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Specifications
Unfortunately,
even today, there is no LC detector that has an equivalent performance to the
flame ionization detector (FID) used in GC. In general, LC detectors have
sensitivities of two to three orders of magnitude less than their GC
counterparts and linear dynamic ranges one to two orders of magnitude lower.
Only highly specific LC detectors have sensitivities that can approach those of
GC detectors.
See also the section on detectors in the HPLC supplement.
Detector Specifications
Detector
specifications are like those for GC detectors and are listed as follows,
1. Dynamic Range
2.
HPLC-Detectors Specifications
Author: RPW Scott
Book:Liquid Chromatography
Section:HPLC UV-Detectors
Detectors
A large number
of LC detectors have been developed over the past thirty years based on a
variety of different sensing principles. However, only about twelve of them can
be used effectively for LC analyses and, of those twelve, only four are in
common use. The four dominant detectors used in LC analysis are the UV
detector (fixed and variable wavelength) the electrical conductivity
detector, the fluorescence detector and the refractive index
detector. These detectors are employed in over 95% of all LC analytical
applications. These four detectors will be described and for those readers
requiring more information on detectors are referred to
Liquid Chromatography Detectors. The subject of detector specifications will not be
discussed here but will also be dealt with in detail there. Detector
sensitivities and detector linearity will, however, be given for each of the
four detectors.
The UV Detector
The UV
detector is by far the most
HPLC UV-Detectors
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Introduction
the lack of an inline
sensitive detector. The first, effective inline liquid chromatography (LC)
detectors were the refractive index detector reported by Tiselius and Claesson
(1) in 1942 and the conductivity detector described by Martin and Randall (2)
in 1951. These two devices should have evoked a growth in LC development, but,
in the early fifties, gas chromatography (GC) was invented which completely
eclipsed the development of LC. It was not until the early 1960s that the
renaissance of LC took place, initially based on the use of the refractive
index of Tiselius and Claesson. Although a significant number of GC detectors
were developed over two or three years, the development of LC detectors was
much slower, largely due to the fact that low concentrations of solute in a
liquid do not change the properties of a liquid nearly as much as they do a
gas. In fact, the development of LC detectors was gradual and arduous.
In a similar
way to the development of GC there has been a
HPLC-Detectors Introduction
Author: RPW Scott
Book:Gas Chromatography
Section:YES Detectors Flame-Ionization
GC Detectors
A large number
of GC detectors have been developed and made commercially available. In
general, GC detectors are 4 to 5 orders of magnitude more sensitive than LC
detectors and, thus, are ideal for trace analysis and environmental monitoring.
The detectors with the highest sensitivity tend to be specific and sense
specific types of sample (e.g., halogenated substances by the electron
capture detector). Conversely, those detectors with a catholic response,
although highly sensitive compared to LC detectors (e.g. the flame
ionization detector) are significantly less sensitive than the specific
detectors. The detectors with a catholic response are the most popular and the
majority of GC separations are monitored by the flame ionization detector
(FID). The most commonly used specific detectors are the nitrogen phosphorus
detector (NPD) and the electron capture detector (ECD) The katharometer
detector,
YES Detectors Flame-Ionization
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Refractive-Index-Based
LC Detectors Based on Refractive Index Measurement
LC detectors
range from those that are exclusively non specific (i.e., bulk property
detectors, e.g., the refractive index detector) through those that are
partially specific (i.e. partial solute property detectors, e.g.,
the UV detectors) to the totally specific detectors (i.e.,
solute property detectors, e.g., the fluorescence detector). In general,
the sensitivity increases progressively as the detector becomes more specific, the
highest sensitivities being obtained from the specific detectors.
Refractive
index is a bulk property of the column eluent and so detection depends on the
solute modifying the overall refractive index of the mobile phase sufficiently
to provide a signal twice that of the noise. Bulk property detectors have an
HPLC-Detectors Refractive-Index-Based
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Time-Constant
and sensor
volumes of the detector. Dispersion of this type is particularly serious in LC
where solute diffusivities are 4 to 5 orders of magnitude smaller than those in
gasses. In GC however, due to the much higher diffusion rates detector
dispersion is minimal and does not significantly effect chromatographic
performance. Consequently detector dispersion in GC detectors will not be
discussed in this book, but dispersion in LC detectors will be considered in
detail in Liquid Chromatography Detectors
.
Peak Dispersion from the Overall Detector Time Constant
Peak dispersion resulting from the time
constant of the sensor and its associated electronics can be significant in
both GC and LC, particularly when filter circuits are introduced to remove
inherent detector noise. The effect of the detector time constant can be
theoretically examined (see
Extra Column Dispersion ) and calculated and
the results from such calculations are shown in figure 6. The undistorted peak,
that
GC-Detectors Time-Constant