Injection
A sample can be placed on a chromatographic column in a variety of ways but only two methods are in general use, either by the use of a syringe or by the use of a sample valve. When a syringe is employed, it is said that the sample has been placed on the column by injection or has been injected onto the column. The sample syringe is calibrated and can have a range of capacities e.g. 1, 5, 10 or 20 microliters. In a packed column, the injection may be made directly into the packing or into a flash hater that volatilizes the sample before it enters the column. In a capillary column the sample is either injected into a flash heater followed by a ’splitter’ or when different types of sample focusing are employed, then a wide bore capillary is used and the sample is place directly into the column. An injection procedure is often used to fill a sample loop prior to placing the sample on the column but in this case, it is not the syringe that is actually placing the sample on the column.
Author: RPW Scott
Book:Preparative Chromatography
Section:Preparative Mass-Overload
nbsp;
Examination of
figure 6 shows that the descending steps of the curves, which correspond to the
tails of the normal elution curves, are very diffuse compared with the ascending
steps, which correspond to the fronts of the normal elution curves. If the
sample is placed on the column by allowing the mobile phase to flow through the
loop for a given time (determined from the flow rate and the volume of sample
selected for injection) and the valve then rotated to allow the mobile phase to
pass directly to the column, the dispersion effect of the sample tube is
virtually eliminated. The improvement resulting from this technique (often
called back-cutting injection) is shown
in the upper curve in figure 6. It is seen that the descending steps of the
curve are very similar to the ascending steps showing that the tube dispersion
has been significantly reduced. It should be pointed out that this injection
procedure
Preparative Mass-Overload
Author: RPW Scott
Book:Gas Chromatography
Section:YES Injection-Devices Open-Tubular-Column
The glass
liner can be fitted with a separate heater and the volatalization temperature
can, thus, be controlled. This "flash heater" system is available in
most chromatographs. By using a syringe with a long needle, the tip can be made
to penetrate past the liner and discharge its contents directly into the column
packing. This procedure is called 'on-column injection' and, as it reduces peak
dispersion on injection and thus, provides higher column efficiencies, is often
the preferred procedure.
Open Tubular Column Injection Systems
Due to the
very small sample size that must be placed on narrow bore capillary columns, a
split injection system is necessary, a diagram of which is shown in figure 8.
Figure 8 The Split
Injection System
The basic
difference between the two types of injection systems is that the capillary
column now projects into the
YES Injection-Devices Open-Tubular-Column
Author: RPW Scott
Book:Capillary Chromatography
Section:Capillary Apparatus Injection-Systems
diameter open tubular columns have been employed that would permit on-column injection. The columns have an I.D. of about 0.056 in., which is slightly greater than the diameter of a specific hypodermic needle. The injection system is shown in figure 7.
Figure 7. Device for On-Column Injection in Large Bore Capillary Columns
Unhappily, this type of injector also is far from ideal, not so much from poor accuracy and precision but from its effect on column resolution. On injection, the sample breaks up into discrete segments, due to bubble formation in the first part of the column. As the solvent evaporates the sample is deposited at two or more locations along the column. When development commences, each local concentration of sample acts as a unique injection and a chromatogram containing very wide or multiple peaks is produced. There have been a number of procedures introduced in an attempt to eliminate the sample splitting on the column. The first solution
Capillary Apparatus Injection-Systems
Author: RPW Scott
Book:Gas Chromatography
Section:YES Injection-Devices Open-Tubular-Column
of the sample.
This problem
was partially solved by using larger diameter columns that would permit on-column
injection. The columns are constructed to have an I.D. of about 0.056 in; which
is slightly greater than the diameter of a certain hypodermic needles. This
injection system is depicted in figure 9.
Figure 9
On-Column Injector for Large Bore Open Tubular Columns
However, there
are also difficulties associated with this type of injector. On injection, the
sample breaks up into separate portions, and bubbles form at the beginning of
the column causing the sample to be deposited at different positions along the
open tube as the solvent evaporates. On starting to develop the separation,
each local concentration of sample acts as a separate injection. As a
consequence, a chromatogram containing very wide or multiple peaks is produced.
Procedures have been introduced in an attempt to eliminate sample splitting in
this manner
YES Injection-Devices Open-Tubular-Column
Author: RPW Scott
Book:Principles and Practice of Chromatography
Section:Principles Nomenclature
has evolved over the years but it was not until the late 1950s
that the various terms used for the characteristics of a chromatogram were
rationalized. A summary of the nomenclature is shown diagramatically in figure
4. The various terms are defined as follows.
Figure 4 The Nomenclature of
a Chromatogram.
The baseline
is any part of the chromatogram where only mobile phase is emerging from the
column.
The peak maximum is the
highest point of the peak.
The injection
point is that point in time/position time when/where the sample is placed
on the column.
The dead
point is the position of the peak-maximum of an unretained solute.
The dead
time (to) is the time elapsed between the injection point
and the dead point.
The dead
volume (Vo) is the volume of mobile phase passed through the
column between the injection point and the dead point
Principles Nomenclature
Author: RPW Scott
Book:Capillary Chromatography
Section:Capillary Apparatus Injection-Systems
nbsp;
Injection Systems
Due to the small dimensions of the column, a very small mass must be injected and this must take the form of a very sharp band of solute entering the column. There are a number of different injection devices that can be used and the appropriate form will depend on the dimensions of the capillary column, in particular its internal diameter. For columns that have diameters that preclude the entry of a syringe needle, a split injector must be used, a diagram of which is
Capillary Apparatus Injection-Systems