Hydrogen
Hydrogen is diatomic gas having a molecular weight of 2 and is used in chromatography, almost exclusively, to provide the combustion gas in a flame ionization detector (FID). In the early days of gas chromatography hydrogen was also used as a carrier gas and the carrier gas was burnt directly at a small jet in both the flame thermocouple detector and the FID. The use of hydrogen as a carrier gas also improved column efficiency. However, hydrogen as a carrier gas has two disadvantages, one its inflammability and two, its tendency to hydrogenate labile substances during their passage through the column. Hydrogen is not a desirable gas to have piped around a laboratory but unfortunately the very popular FID detector and the NPD detector demand its use. As a result hydrogen generators have found popularity, as they produce hydrogen by electrolysis as required and, thus, significantly reduce any fire risk. The advantages in column performance that is provided by using hydrogen as a carrier gas can be partially realized by using helium as the carrier gas.
Author: RPW Scott
Book:Gas Chromatography
Section:YES Gas-Supplies Hydrogen-Generators
at pressures from 60 to 100 psi.
Hydrogen Generators
In the Packard
Hydrogen Generator, hydrogen is generated electrolytically from pure deionized
water. Unfortunately, the technology used in hydrogen generators is largely
proprietary and technical details are not readily available. The electrolysis
unit uses a solid polymer electrolyte and thus does not need to be supplied
with electrolytes, only the deionized water. The manufacturers claim the device
generates 99.999% pure hydrogen with a reservoir capacity of 4 liter, and an
output pressure that ranges from 2 to 100 psi. Other units can produce hydrogen
flows that range from 0 to 125 ml/min. to 0 to 1200 ml/min. The oxygen,
produced simultaneously with hydrogen at half the flow rate, is vented to air
YES Gas-Supplies Hydrogen-Generators
Author: RPW Scott
Book:Liquid Chromatography
Section:HPLC Stationary-Phases Silica-Gel Structure
be hydrogen bonded to the hydroxyl
groups and multi-layers of water physically adsorbed on top of these. Water can
be hydrogen bonded to the silica gel surface in a number of different ways
which are depicted in figure 31.
None of
the above structures has been confirmed in an unambiguous manner but all are
reasonably possible. The center and right hand side structures contain a type
of double hydrogen bond and would have high energies of formation and, thus,
more stable than the simple hydrogen bond depicted on the left. The right hand
structure would be particularly stable as it constitutes a four membered
hydrogen bonded ring similar to that which might be expected to form in the
strong association of water with itself.
Figure 31. Different
Ways in Which Water May be Hydrogen Bonded to Silica Gel Hydroxyl
Groups
HPLC Stationary-Phases Silica-Gel Structure
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Sensor-Volume
a hydrogen, or hydrogen/nitrogen stream, which
is then burnt at a small jet and the ions produced measured by an appropriate
pair of electrodes. The response of the detector will depend on the mass of
solute eluted per unit time from the capillary column and, thus, will be
independent of the hydrogen or hydrogen/nitrogen flow (as this will not effect
the rate of solute elution, in terms of mass per unit time). In contrast, if
the FID was a concentration device (like the katherometer), as the hydrogen or
hydrogen/nitrogen flow dilutes the sample, the response will be directly
related to the flow rate of the diluting gas. It follows, that for a mass
sensitive device, the sensing volume can be extremely mall and, in fact,
insignificant. However, for a device to sense a concentration change, the
sensor must have a finite (albeit small) volume
EC-Dispersion Sensor-Volume
Author: RPW Scott
Book:Capillary Chromatography
Section:Capillary Detectors The-Nitrogen-Phosphorus-Detector
14. The sensor of the NPD is a small rubidium or cesium bead contained inside a small heater coil. The helium carrier gas is mixed with hydrogen and passes into the detector through a small jet. The bead, which is heated by passing a current through the coil, is situated above the jet, and the helium-hydrogen mixture (produced by mixing the column carrier gas, helium with a separate stream of hydrogen) passes over it. If the detector is to respond to both nitrogen and phosphorus, then a minimum hydrogen flow is employed to ensure that the gas does not ignite at the jet. In contrast, if the detector is to respond to phosphorus only, a large flow of hydrogen can be used and the mixture burned at the jet. A potential is applied between the bead and the anode. The heated alkali bead emits electrons by thermionic emission which are collected at the anode and thus produce an ion current. When a nitrogen or phosphorus containing solute is eluted, the partially combusted nitrogen and
Capillary Detectors The-Nitrogen-Phosphorus-Detector
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Transport Modified-Moving-Wire
to a 2 in. length of 1/2 in I.D. thin walled stainless steel tube
filled with about 2 g of nickel catalyst and then to the aspirator. The wide
tube was closed with a loose plug of quartz wool and .
Figure 43 The Modified Moving
Wire Detector
The nickel
catalyst was prepared by absorbing a saturated solution of nickel nitrate onto
20/40 BS mesh brick dust, decomposing the nitrate at 500˚C for 3 hours
followed by reduction of the nickel oxide so produced to metal in a stream of
hydrogen at 250˚C. The jet/venturi aspirator (supplied commercially as a molecular
entrainer) was placed in line with the hydrogen flow to the detector.
To improve the
aspirating efficiency, the gas used was a mixture of hydrogen and argon, and
with this mixture, the jet/venturi pressure drop continuously sucked the
combustion gasses into the hydrogen stream. In figure 43 the reduced pressure
side of the aspirator is shown connected to the side limb of the oxidation tube
and the two
HPLC-Detectors Transport Modified-Moving-Wire
Author: RPW Scott
Book:Gas Chromatography
Section:YES Detectors Nitrogen-Phosphorus-(NPD)
The actual NPD
sensor is a rubidium or cesium bead contained inside a small heater coil. The
helium carrier gas is mixed with hydrogen and passes into the detector through
a small jet. The bead is heated by a current passing through the coil which is
situated above the jet, and the helium-hydrogen mixture passes over it. If the
detector is to respond to both nitrogen and phosphorus, then a minimum hydrogen
flow is employed to ensure that the gas does not ignite at the jet. In
contrast, if the detector is to respond to phosphorus only, a large flow of hydrogen
can be used and the mixture burned at the jet. A potential is applied between
the bead and the anode. The heated alkali bead emits electrons by thermionic
emission which are collected at the anode and thus produce an
YES Detectors Nitrogen-Phosphorus-(NPD)