Gradient
Gradient elution is used in LC for the same purpose that temperature programming is used in GC, that is, to accelerate the elution of strongly retained solutes. Depending on the interactive character of the stationary phase, the interactive character of the mobile phase is continually changed to have interactive properties more like those of the stationary phase during the development of a separation. As an example, if a reversed phase (e.g. silica bonded with long hydrocarbon chains) was used as the stationary phase in a liquid chromatography separation, the solutes would be held and retained predominantly by dispersion forces. To allow the dispersion forces to dominate in the separation process and retain the solutes, the mobile phase would be made strongly polar, for example, a 10% solution of acetonitrile in water. Now the solvent that has the greater dispersive character in the mobile phase mixture is acetonitrile. Thus, if the proportion of acetonitile is increased during chromatographic development, the dispersive interactions in the mobile phase will become larger until they can compete with those of the stationary phase and elute the dispersively held solutes. Conversely, if the stationary phase was polar (e.g. silica gel) the mobile phase would be made predominantly dispersive (e.g. 1% ethyl alcohol in n-heptane). The mobile phase would then be made progressively more polar by increasing the ethyl alcohol content, so that the polar interactions between the ethyl alcohol and the solutes would be able to compete with the polar interactions of the solutes with the hydroxyl groups of the silica gel and allow them to be eluted more quickly. There is a wide range of solvent mixtures that can be used in gradient elution and the two examples given are extreme to illustrate the basic factors controlling gradient elution. In practice, were the interactive difference between solutes is very subtle, very subtle mixtures of solvents may be necessary to make the gradient elution process successful.
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Rapid-Separations
use.
Rapid Separations Employing Gradient Elution
Rapid LC
separations are relatively easy to accomplish with isocratic development
assuming low dispersion instrumentation with a fast response is available.
However, if a sample mixture contains components that extend over a wide
polarity or molecular weight range then gradient elution development will be
necessary and fast gradients are almost impossible to form with conventional LC
solvent programmers. As a consequence, for high speed gradient separations, a
unique procedure must be used in conjunction with specially designed apparatus.
The solution to
the problem of fast gradient generation is to employ a preformed gradient a
concept that was first introduced by Snyder and Saunders (13) as long ago as
1969. A diagram of a gradient preformer is shown in figure 27. In the
particular apparatus that was used to provide the fast analyses that are
described below the required gradient was formed in a column 25 cm long, 4.6 mm
EC-Dispersion Rapid-Separations
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Rapid-Separations
to provide a solvent
concentration profile is formed over a period of time and pumped into the
gradient storage column. During the process of loading the gradient into the
storage column, the solvent content of the storage vessel is passed to waste.
When the complete solvent program is contained in the storage column, the flow
is arrested. The sample is then charged into the sample loop (an internal
sample valve loop should be used). The loop is then placed in line with the
column and the gradient is discharged at full flow rate through the sample loop
and column. An example of the rapid separation of a thirteen component mixture
in just over 20 seconds is shown in figure 28.
J. Chromatogr.,253(1982)159
Column Length 2.5 cm, column I.D. 2.6 mm,
packing, C18 reversed phase, particle size 3 mm, solvent program 25% v/v acetonitrile in
water to 100 % acetonitrile, flow rate 5 ml/min.
Figure 28.
The fast Separation of a Wide-Polarity Range Mixture by Preformed Gradient
EC-Dispersion Rapid-Separations
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Rapid-Separations
Although the
actual elution took only 22 seconds, due to the time required to form the
gradient and regenerate the column the total gradient cycle was 5-6 minutes. To
fully utilize the speed of the system a number of gradient storage columns
would be necessary, that could be operated in parallel, if the gradient
analysis was to be repeated continuously. The quantitative repeatability of the
system was tested with 8 replicate analyses of the mixture. The results
obtained are shown in Table 6. It is seen that despite the complexity of the
analytical procedure, and the need for
EC-Dispersion Rapid-Separations
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Transport Modified-Moving-Wire
nbsp;
Figure
44 The Separation of Blood Liquids Employing Incremental Gradient Elution
and Monitored by the Modified Moving Wire Detector
However, due
to the limited number of compounds that were tested this relationship should be
assumed only with caution. A chromatogram of blood lipids obtained by
incremental gradient elution and monitored by the modified detector is shown in
figure 44. As incremental gradient elution involves a program of 12 solvents
ranging from hydrocarbons, chlorinated
hydrocarbons, nitro-paraffins,
esters, ketones and alcohols. This
HPLC-Detectors Transport Modified-Moving-Wire
Author: RPW Scott
Book:Liquid Chromatography
Section:HPLC HPLC-Mobile-Phases Aqueous-Solvent-Mixtures
of the unique characteristics of mobile phases
consisting of methanol water mixtures when used in reversed phase LC. From
figure 47 it is seen that when the original mixture contains 50%v/v of methanol
there is little free methanol available in the mobile phase to elute the
solutes as it is mostly associated with water. Subsequently, however, the
amount of methanol unassociated with water increases rapidly in the solvent
mixture and this rapid increase must be accommodated by the use of a convexgradient profile when employing gradient elution. The convex gradient
will compensate for the strongly concave form of the unassociated
methanol concentration profile shown in figure 47 which will be the strongest
eluting component of the mobile phase. The strong association of methanol with
water could also account for the fact that proteins can tolerate a significant
amount of methanol in the mobile phase before they become denatured. It is
clear that this is because there is virtually no
HPLC HPLC-Mobile-Phases Aqueous-Solvent-Mixtures
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Electrochemical
and the
reference electrode (which compensates for any changes in the background
conductivity of the mobile phase).
The processes
taking place at the electrode surface can be very complex; nevertheless, the
dominant reaction can be broadly described as follows.
At the actual
electrode surface the reaction is extremely rapid and proceeds almost to
completion. This results in the layer close to the electrode being virtually
depleted of reactant. As a consequence, a concentration gradient is established
between the electrode surface and the bulk of the solution.
This
concentration gradient causes solute to diffuse into the depleted zone at a
rate proportional to the solute concentration in the bulk of the mobile phase.
Thus, the current generated at the electrode surface will be determined by the
rate at which the solute reaches the electrode and consequently, as the process
is diffusion controlled, will depend on solute concentration and the magnitude
of solute
HPLC-Detectors Electrochemical