411
Fluorescence
Fluorescence is a specific type of luminescence. If molecules are excited by electromagnetic radiation to produce luminescence, the phenomenon is termed photoluminescence. If the release of radiation is immediate, or stops on the removal of the exciting radiation, the substance is said to be fluorescent. If, however, the release of energy is delayed, or persists after the removal of the exciting radiation, then the substance is said to be phosphorescent. Due to its persistence, phosphorescence is of little use in chromatography detection, but fluorescence has been used to produce some of the most sensitive LC detectors (albeit somewhat specific in their response). When light is adsorbed by a molecule, a transition to a higher electronic state takes place and this absorption is highly specific to the substance concerned. Radiation of a particular energy (or wavelength) is only absorbed by a specific structure. The adsorption of light energy produces the characteristic absorption spectrum of a molecule. If the absorbed energy is not rapidly dissipated by collisions with other molecules, or by other means, the electron will return to its ground state, with the emission of electromagnetic energy producing fluorescence. As some energy is always lost in the process, the fluorescent light always has a longer wavelength than the excitation light.
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Fluorescence Multi-Wavelength-Fluorescence
The Multi Wavelength Fluorescence Detector
The multi
wavelength fluorescence detector contains two monochromators, one to select the
excitation wavelength and the second to select the fluorescence wavelength or
produce a fluorescence spectrum A diagram of the multi wavelength fluorescence
detector is shown in figure 38.
Figure 38. The
Fluorescence Spectrometer Detector
The detector
comprises a fluorescent spectrometer fitted with suitable absorption cell that
is sufficiently small so as not to degrade the resolution of an LC column. There are
HPLC-Detectors Fluorescence Multi-Wavelength-Fluorescence
Author: RPW Scott
Book:Gas Chromatography - Tandem Techniques
Section:GC-Tandem Fluorescence-Spectrometer
fluorescence spectrometer must be fitted with a sensor cell of appropriate dimensions. Such a spectrometer system can be highly complex and versatile and allows excitation spectra to be obtained at any fixed fluorescent wavelength or emission spectra to be obtained for any fixed excitation wavelength. In general fluorescence spectra have very limited use in structure elucidation but can be used for identification purposes providing the necessary reference spectra are available. A diagram of a fluorescence spectrometer is shown in figure 13. It consists of two monochromators, the first that selects the wavelength of the excitation light and the second disperses the fluorescent light and provides a fluorescence spectrum. The spectrometer incorporates two distinctly different light paths and as a result the optical system appears quite complicated. If the different light paths are considered separately, that is firstly, the path of the excitation light and secondly, the path of the
GC-Tandem Fluorescence-Spectrometer
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Fluorescence Single-Wavelength-Excitation
9. Chrysene
10. Benzo(b)fluoranthene
11. Benzo(k)fluoranthene
12. Benzo(k)fluoranthene
13. Dibenz(a,h)anthracene
14.Indeno(1,2,3,cd)pyrene
15. Benzo(ghi)perylene
Courtesy of the Perkin Elmer Corporation
Figure 37.
Separation of the Priority Pollutants Monitored by the Simple Fluorescence
Detector
There are some
compromises between the expensive fluorescence spectrometer detector and the
single wavelength excitation fluorescence detector. Some have a single
monochromators that select the wavelength of the excitation light, others
employ a single monochromator to select the emission wavelength or provide
emission spectra
HPLC-Detectors Fluorescence Single-Wavelength-Excitation
Author: RPW Scott
Book:Liquid Chromatography
Section:HPLC Refractive-Index
lamp. This simple type of fluorescence
detector was the first to be developed, it is relatively inexpensive and for
certain compounds can be extremely sensitive. Typical specifications for a
fluorescence detector are as follows:-
Typical
Specifications for a Fluorescence Detector
Sensitivity (Anthracene)
1x 10-9 g/ml
Linear Dynamic Range
1 x 10-9 to 5 x 10-6 g/ml
Response Index
0.96 - 1.04
A more
elaborate form of fluorescence detector uses a monochromator to select the
excitation wavelength and a second monochromator to select the wavelength of
the fluorescent light. This instrument gives the maximum versatility and allows
the maximum sensitivity to be realized for any type of solute. The system can
also provide a fluorescence spectra by arresting the flow of mobile phase when
the solute resides in the detecting cell and scanning the fluorescent light.
The Refractive Index Detector
The
HPLC Refractive-Index
Author: RPW Scott
Book:Gas Chromatography - Tandem Techniques
Section:GC-Tandem GC-Spectroscopic-Systems
reference compound and, if a match is obtained (the difference is close to zero) the unknown is considered identified. In a similar manner, for certain compounds, UV and fluorescence spectra can also be used to confirm solute identity and the association of the GC with a UV spectrometer or a fluorescence spectrometer has proven to be much easier due to the significantly greater sensitivity of these spectrometers compared with that of the IR spectrometer.
The use of UV spectra and fluorescence spectra, however, are far less useful than mass spectra or IR spectra for solute identification. Except for certain substances (e.g. those containing aromatic rings) the majority of compounds give very similar UV spectra with very little fine structure to allow confident spectra matching. This is due to a multiplicity of adsorption bands merging to produce a broad envelope the cause of which will be discussed later. Similarly fluorescence spectra have much less detail than IR
GC-Tandem GC-Spectroscopic-Systems
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Fluorescence
The Fluorescence Detector
When light is
emitted by molecules that are excited by electromagnetic radiation, the
phenomenon is termed photoluminescence. If the release of electro-magnetic energy
is immediate, or stops on the removal of the
excitation radiation, the substance is said to be fluorescent. If, however, the
release of energy is delayed, or
persists after the removal of the exciting radiation, then the substance is
said to be phosphorescent. Fluorescence has been shown to be extremely
HPLC-Detectors Fluorescence