Flame Ionization Detector (FID)
The FID is predominantly a GC detector although it has been used as a sensor in special types of LC detectors (e.g., the transport detector). The FID evolved from the flame thermocouple detector in which a stream of hydrogen is burnt at a small jet over which a thermocouple is situated. The mobile phase from the GC column is fed into the hydrogen stream and when a solute is eluted, the temperature of the flame increases together with the thermocouple output. The thermocouple output is monitored by a backing-off circuit and a potentiometric recorder. In the FID, the thermocouple is replaced by an electrode situated away from the flame and a potential of a 100-200 volts applied between the jet and the electrode. This potential collects any ions or electrons formed in the flame during combustion of an eluted solute which is monitored as an increase in ion current by a high impedance amplifier. During the elution of a carbon containing substance, it is thought that microscopic particles of carbon are formed that thermally emit electrons and which provide the signal current to the high impedance amplifier. Although the signal current is very small (the ionization efficiency is only 0.0015%) the noise level is also very small (ca <10-13 amp) and, with a well optimized system, sensitivities of 5 x 10-12 g/ml for n-heptane at a signal/noise ratio of 2 can be easily realized. Except for a very few organic compounds (e.g. carbon monoxide, etc.) the FID detects all carbon containing compounds. The detector also has an extremely wide linear dynamic range that extends over, at least, five orders of magnitude with a response index between 0.98-1.02.
Author: RPW Scott
Book:Gas Chromatography
Section:YES Detectors Flame-Ionization
.
The detectors with the highest sensitivity tend to be specific and sense
specific types of sample (e.g., halogenated substances by the electron
capture detector). Conversely, those detectors with a catholic response,
although highly sensitive compared to LC detectors (e.g. the flame
ionization detector) are significantly less sensitive than the specific
detectors. The detectors with a catholic response are the most popular and the
majority of GC separations are monitored by the flame ionization detector
(FID). The most commonly used specific detectors are the nitrogen phosphorus
detector (NPD) and the electron capture detector (ECD) The katharometer
detector, although having relatively poor sensitivity is widely used in gas
analysis.
The Flame Ionization Detector
The FID,
invented by Harley and Pretorious (7), and separately by McWilliams and Dewer
(8), evolved from the Heat of Combustion Detector developed by Scott (9). The FID detector employs
YES Detectors Flame-Ionization
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Flame-Ionization
.
The Flame Ionization detector
Without doubt, the Flame Ionization Detector (FID) is the most useful GC detector
available and by far that most commonly used in GC analyses. The FID has a very
wide dynamic range, a high sensitivity and (with the exception of a few low
molecular weight compounds) will detect all substances that contain carbon. The
first FID was described about the same time by Harley and Pretorious (12), and
McWilliams and Dewer (13). The FID is an extension of the flame thermocouple
detector and is physically very similar, the fundamentally important difference
being that the ions produced in the flame are measured as opposed to the heat
generated. Hydrogen is mixed with the column eluent and burned at a small jet.
Surrounding the flame is a cylindrical electrode and a relatively high voltage
is applied between the jet and the electrode to collect the ions that are
formed in the flame. The resulting current is amplified by a high impedance
amplifier and
GC-Detectors Flame-Ionization
Author: RPW Scott
Book:Capillary Chromatography
Section:Capillary Detectors Flame-Ionization-Detector
for use with capillary columns. The nitrogen phosphorous detector (NPD) is also appropriate for capillary columns (the function of which is very similar to that of the FID) for the same reason. The micro-argon detector employs a scavenger flow which also, in effect, reduces the sensor volume, so this detector can also be used with capillary columns. The FID and the NPD detector will be described here. For a more detailed treatment of GC detectors see book 4 of this series.
The Flame Ionization Detector
The FID, invented by Harley and Pretorious (6), and separately by McWilliams and Dewer (7), evolved from the Heat of Combustion Detector developed by Scott (8). The FID operates in the following manner. Hydrogen is mixed with the column exit gas (which may be helium, hydrogen or any other appropriate gas) and then burnt at a small jet which is situated inside a cylindrical electrode system.
Figure 13. The Flame Ionization Detector
Capillary Detectors Flame-Ionization-Detector
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Early-GC-Detectors Flame-Thermocouple
capacity and thermal conductivity from those of the
carrier gas. This detector was used extensively in the early days of GC for the
analysis of hydrocarbon gases. There was much discourse and dissent with
regards to the exact mechanism of detection involved in the katharometer and
even today it is considered to respond to a number of different physical
properties of the eluent gas with no one property playing a major role.
The Flame Thermocouple Detector
The "flame thermocouple detector"
was the next detector to be reported which was developed by Scott [8] and was,
in fact, the forerunner to the flame ionization detector FID. A diagram of the
flame thermocouple is shown in figure 8.
Figure
8 The Flame Thermocouple Detector
GC-Detectors Early-GC-Detectors Flame-Thermocouple
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Ionization-Detectors Electron-Capture
1 x 10-9
to 5 x 10-9 amp, the noise level is about 1.2 x 10-13 amp and the ionization efficiency is
about 0.07%. It is claimed to be about
10 times more sensitive than the flame ionization detector and to have a linear
dynamic range of 105. An example of the use of a pulsed helium
discharge detector for monitoring the separation of some aromatics on a
capillary column is shown in figure 37. The pulsed helium discharge detector
appears to be an attractive alternative to the flame ionization detector and
would eliminate the need for three different gas supplies. It does, however,
require equipment to provide specially purified helium, which diminishes the
advantage of using a single gas.
The Electron Capture Detector
Lovelock's work on ionization detectors
culminated in the invention of the electron
capture detector (25). However,
the electron capture detector operates on an entirely different principle
from that of the argon detector. A low energy b-ray
source is used
GC-Detectors Ionization-Detectors Electron-Capture
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Early-GC-Detectors Emissivity
90zirconium.
The ionization currents formed (collected by
appropriate electrode potentials) are arranged to oppose one another and
consequently any variation in pressure or temperature of the two cells will be
balanced out. The differential signal resulting from the presence of a solute
in the column eluent sensor cell is amplified and recorded. The sensitivity of
the detector was similar to that of the katharometer i.e. about 1 x 10-6
g/ml. Unfortunately, the practical lifetime of this detector was also
relatively short as it was eclipsed by both the FID and the argon ionization
family of detectors subsequently introduced by Lovelock.
The Emissivity Detector
Another interesting
early detector was the emissivity detector developed by Grant. an interesting
and innovative extension of the flame thermocouple detector.
 
GC-Detectors Early-GC-Detectors Emissivity