Efficiency
The efficiency of a chromatographic column is a measure of the capacity of the column to restrain peak dispersion and thus, provide high resolution. The higher the efficiency, the more the peak dispersion is restrained, and the better the column. The column efficiency will vary with the retention of the peak. In capillary columns, the efficiency generally falls as the retention increases and for a packed column the efficiency generally increases with retention. The expression for calculating the column efficiency can be derived from the plate theory. Column efficiency is measured in theoretical plates (from the Plate Theory) and is taken as 16 times the square of the ratio of the retention distance (the distance between the injection point and the peak maximum) to the peak width at the points of inflection (the points of inflection occur at 0.6065 of the peak height). The Height Equivalent to the Theoretical Plate (HETP) or the Variance per unit Length of a Column is calculated as the ratio of the column length to the column efficiency.
Author: RPW Scott
Book:Principles and Practice of Chromatography
Section:Principles Peak-Dispersion TLC-Plates
of the silica layer must be
made as small as possible and the layer must be spread in a thin, homogenous
film on the supporting plate. TLC plate efficiency is measured in a similar
manner to column efficiency but slightly modified. It is very difficult, if not
impossible, to identify the positions of the points of inflexion on a TLC spot,
but if the visible edges of the spot are assumed to occur at four standard
deviations of the spot distribution, then it is still possible to assess the efficiency. In general it is considered that over 95% of the material in the
spot is confined within 4 standard deviations of the spot dispersion.
If the
diameter of the spot (d), corresponds to four standard deviations, then
applying the same rationale as with the packed column,
where (Zs)
is the retention distance of the solute.
Thus,
It shout be
pointed out, however, the method contains
Principles Peak-Dispersion TLC-Plates
Author: RPW Scott
Book:The Thermodynamics of Chromatography
Section:Thermodynamics Other-Methods Chiral-Separations Column-Efficiency
nbsp
The Effect of Temperature and Solvent Composition on the Required Column Efficiency
Using the
values for the capacity ratios and separation ratios derived from equations
(47), (48) and (49) in equation (39) the efficiency necessary to ensure a
separation of (6s) for the two
enantiomers can be calculated over a range of temperatures and solvent
compositions.
Figure 22.
Graphs of Required Efficiency against Temperature for Each Solvent Composition
Curves
relating required efficiency against temperature for each solvent composition,
calculated in this manner, are shown in figure 22. As would be expected, the
minimum efficiency is required at the lowest temperature and lowest ethanol
concentration. As either the separation ratio and/or the capacity ratios
decrease, the necessary efficiency to achieve a separation increases (as
predicted by equation (39)). At one extreme, where the capacity ratio is very
small (i.e. at 50% v/v ethanol and 50˚C), 15000
Thermodynamics Other-Methods Chiral-Separations Column-Efficiency
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Van-Deemter-Equation
efficiency attainable at a given pressure. This is
because, as the particle diameter is increased the column permeability is also
increased allowing a longer column to be used. The permeability increases as
the square of the particle diameter but the variance per unit length only
increases linearly with the particle diameter. Thus, doubling the particle
diameter will allow a column four times the length to be used but the number of
plates per unit length will be halved. Consequently, the columnefficiency will be increased by a factor of two. It is also seen that the
higher efficiencies will be obtained with mobile phases of low viscosity and
for solutes of low diffusivity. Solvent viscosity and solute diffusivity tend
to be inversely proportional to each other and so the sensitivity of the
maximum obtainable efficiency to either solvent viscosity or solute diffusivity
will generally not be large. The approximate length of a column that will
provide the maximum column efficiency when
Dispersion Van-Deemter-Equation
Author: RPW Scott
Book:Preparative Chromatography
Section:Preparative Loading-Capacity
the column
efficiency (unless it is accompanied by an corresponding increase in the
particle diameter).
However,
increasing the column efficiency will have the opposite effect, as seen by
equation (1), it will reduce the maximum sample load. Consequently, if
the necessary efficiency to achieve the required separation has been obtained,
then if the column is lengthened to increase the loading capacity for optimum
performance, either the flow rate will need to be increased to reduce the
efficiency and thus maintain the maximum loading, or the particle size will
need to be increased to reduce the efficiency to its required value. However,
an increased flow rate will also reduce separation time and thus increase
sample throughput. Conversely, the alternative use of larger particles will
increase column permeability and thus the column can be operated at a lower
pressure and be constructed of lighter materials. Again, a an alternative, if the sample is merely a two
component
Preparative Loading-Capacity
Author: RPW Scott
Book:Plate Theory and Extensions
Section:Plate-Theory Resolving-Power
needed at lower (k') values will require longer columns
which will extend the analysis times. To resolve a solute pair with a
separation ratio of 1.02, an efficiency of 360,000 theoretical would be
required if the (k') value was 0.5. GC Capillary columns can provide such
efficiencies but, in LC, such efficiencies would be extremely difficult and
costly to produce. It follows that the phase system should be chosen so that
the closest eluted solutes are not eluted at low (k') values. Less efficiency
will be needed and, thus, shorter columns and consequently, shorter analysis
times will be achieved. At (k') values that exceed 10, the required
efficiency changes little as the capacity ratio increases. Thus, for fast
analyses, the phase system provide a large separation ratio, but the first peak
should elute at a (k') of 10 or more. The phase system should have high
selectivity and retentive capacity so that minimum efficiency is required and
the column can be as short as possible.&
Plate-Theory Resolving-Power
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Van-Deemter-Equation
nbsp;
Figure 12.
Graph of Log. Maximum Efficiency against Particle Diameter
It is seen from
figure 12 that changing the particle diameter from 1 to 20 micron results in an
efficiency change from about 3500 theoretical plates to nearly 1.5 million theoretical plates and furthermore,
this very high efficiency is achieved at an inlet
pressure of only 3000 p.s.i.. It is also seen that the maximum available
efficiency increases as the particle diameter increases. This is because, as
already discussed, if the pressure is limited, in order to increase the column
length to provide more theoretical plates, the permeability of the column must
be increased to allow the optimum mobile phase velocity to be realized. It is
possible to increase the inlet pressure to some extent, but ultimately the
pressure will be limited and the effect of particle diameter will be the same
but at higher efficiency levels
Dispersion Van-Deemter-Equation