Detector Response
Detector response can be described in two ways. The detector response can be defined as the voltage output for unit change in solute concentration flowing through the detector sensor, or, for a mass sensitive detector, the voltage output for unit change in solute mass passing through the detector. Alternatively, detector response can also be defined as the voltage output that would result from unit change in the physical property the detector measured (e.g., refractive index, or heat of combustion etc.). It is obtained experimentally from the output produced by passing a known mass, or concentration of solute through the detector.
Author: RPW Scott
Book:Liquid Chromatography Detectors
Section:HPLC-Detectors Specifications
.
Only highly specific LC detectors have sensitivities that can approach those of
GC detectors.
See also the section on detectors in the HPLC supplement.
Detector Specifications
Detector
specifications are like those for GC detectors and are listed as follows,
1. Dynamic Range
2. Response Index or Linearity
3. Linear Dynamic range
4. Detector Response
5. Detector Noise Level
6. Detector Sensitivity or Minimum Detectable
Concentration
7. Total System Dispersion
8. Sensor Dimensions
9. Detector Time Constant
10. Pressure Sensitivity
11. Flow Sensitivity
12. Operating Temperature Range
In general the
specifications are the same for both GC and LC detectors with the exception of
detector dispersion. Although, detector dispersion has a minimal
HPLC-Detectors Specifications
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Detector
peak height
(or the peak area) using the numerical value of the response index. Thus. in
effect, the useful linear dynamic range of a detector for quantitative purposes
can be significantly extended by employing correction procedures when using the
response index.
It should be pointed out that the logarithmic
dilution method should not be used if the linearity is to be measured by the
method recommended by the E19 committee of the ASTM.
Detector Response
There are two ways of defining detector
response, either as detector output (usually in mv) per unit change in solute
concentration or as the detector output per unit change in the units of
detector measurement (e.g. the
sensitivity of a conductivity detector would be defined in terms of detector
output per unit change in electrical conductivity). The detector response (RD)
is determined by injecting a known mass (m ) onto the column and measuring the
peak height (h) in (mv), then
GC-Detectors Detector
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Summary
range of a detector is that concentration range over which
it will give a concentration dependent output. The units are dimensionless.
2. The Response Index (r) The response
index of detector is a measure of detector linearity and would be unity for a
truly linear detector. In practice the value of (r) should lie between 0.98 and
1.02. If (r) is known, quantitative results can be corrected for any non
linearity.
3. Linear Dynamic Range (DL)
The linear dynamic range of a detector is that concentration range over which
the detector response is linear within defined response index limits. It is also dimensionless and is important
when the components of a mixture cover a wide concentration range.
4. Detector Response (Rc) The
detector response can be defined as the detector output per unit change in
concentration (e.g. volts/g/ml) or, as the detector output
per unit change of physicalpropertybeingmeasured (e.g.fortheFID, volts/gram of carbon/sec). In
GC-Detectors Summary
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Response
detector, the sensitivity would be defined in units of g/sec.
The Form of Detector Response
There are
three different forms of detector response, namely, proportional, differential
and integral. A proportional response is one that is directly related to the
concentration of solute in the mobile phase passing through it. All detectors
with a proportional response are designed to give as near a linear response as
possible. In many detectors, the actual sensor does not give a proportional
response. Thus suitable electronic circuitry must be employed to modify the
signal from the sensor so that the actual detector output is proportional to
the solute concentration in the mobile phase passing through it. For example a
sensor with a logarithmic response would be modified by an exponential
amplifier to give an output linearly related to the solute concentration. The
different types of detector response are shown in figure 1.
Figure 1. Different Types of Detector Response
GC-Detectors Response
Author: RPW Scott
Book:Gas Chromatography Detectors
Section:GC-Detectors Linearity
has a truly linear
response (despite manufacturers claims) but most detectors will have a response
approaching that of linear. It is difficult to apply a standard to detector
linearity, but the Response Index (1)
does help comparisons to be made between one detector with that of another.
Providing the response of the detector
approaches linearity then its response can be described by the following simple
equation,
y = Acr
where (r) is the response index and the other symbols have the
For a truly linear detector, r=1, and the
extent to which (r) deviates from unity would be a measure of its non
linearity.
Curves relating the detector output to
different solute concentrations passing though it for different response values
are show in figure 2.
Figure
2 Curves Relating Detector Output to Solute Concentration for Different
Response Indices
GC-Detectors Linearity
Author: RPW Scott
Book:Principles and Practice of Chromatography
Section:Principles Basic-Chromatograph Output
where (A) is a constant,
(Cm) is the concentration of solute,
and (V) is the output of the detector.
Because of the
imperfections in mechanical and electrical devices practical detectors can only
approach this ideal response. A measure of linearity that is specified in
numerical terms so that comparisons can be made between detectors can be
obtained as follows. It is assumed that for a closely linear detector the
response could be described by the following power function
(1)
where (r) is defined as the Response Index
It follows
that for a truly linear detector, r = 1, and the proximity of (r) to unity will
indicate the extent to which the response of the detector deviates from true
linearity. In addition if (r) is not unity but is known then appropriate
corrections can be made to the
Principles Basic-Chromatograph Output