Compressibility
Fluids are employed as mobile phases in chromatography and all fluids are to some extent compressible. The compressibility of liquids, however, is very small and, thus, has little impact on liquid chromatography performance. Gases, however, have high compressibility, as described by Boyle’s law, which states that at a given temperature, the product of the pressure and volume of a gas is constant. Thus, as there is a significant pressure drop across a GC column, although the mass of gas passing a give point is constant along the column length, the volume of gas passing through the column continually increases from the start of the column to the end. This has important ramifications when attempting to measure the specific retention volume of a solute for thermodynamic analysis. It is clear that, as the volume flow will continually increase as the solute passes along the column, then a pressure correction must be applied. The true retention volume of a solute can be calculated as 1.5 times the product of the retention time and exit flow rate, times (the inlet-outlet pressure ratio squared minus 1) over (the inlet-outlet pressure ratio cubed minus 1).
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Summary
functions contained in it, are well substantiated by
experiment. The Knox equation is obtained from an empirical fit to experimental
data and the individual functions of other pertinent variables contained in the
equation are not all substantiated by experiment. The Golay equation accurately
described dispersion in capillary or open tubular columns but in GC the
compressibility of the mobile phase must also be taken into account (Golay in
his original derivation did not accommodate gas compressibility).
It would
appear from the data available at this time, that the Van Deemter equation (for
packed columns) or the Golay equation (for capillary or open tubular columns)
would be the most appropriate to use in column design and in the interpretation
of column properties
Dispersion Summary
Author: RPW Scott
Book:Principles and Practice of Chromatography
Section:Principles Peak-Dispersion Multi-Path-Effect
Peak Dispersion in a Chromatographic Column
The first
comprehensive approach to dispersion in chromatographic columns was taken by
Van Deemter (8) who developed the dispersion equation for a packed GC column.
Van Deemter's development did not take into account the compressibility of the
mobile phase which was dealt with later by Katz, Ogan and Scott (9). A simple
form of this theory will be given that does not accommodate the compressibility
of the mobile phase but a more detailed and comprehensive treatment is given in
Dispersion in Chromatography Columns
and Extra Column Dispersion.
Van Deemter et
al. assumed that there were four band spreading processes responsible for
peak dispersion, namely, multi-path dispersion, longitudinal
Principles Peak-Dispersion Multi-Path-Effect
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Compressibility-Effects-GC
Effect of Mobile Phase Compressibility On the HETP Equation for a Packed GC Column
As the pressure falls along the column
length, the velocity changes and, as the solute diffusivity depends on the
pressure, the diffusivity of the solute will also change. The multi-path term,
which contains no velocity or gas pressure dependent parameters, will be
unaffected and the expression that describes it the same. The other terms in
the HETP equation, however, all contain parameters that are affected by
gas
Dispersion Compressibility-Effects-GC
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion GC-Capillary-Columns
GC Capillary Columns
The maximum
permissible extra column dispersion for a GC capillary column must be corrected
for the compressibility of the mobile phase and the alternative geometry of the
tube,
Thus, for a GC
capillary tube, equation (3) becomes
and equation
(4) becomes
In a similar
manner, the maximum sample volume that can be placed on a GC capillary column,
assuming half the permissible extra column dispersion is allotted to sample
volume, follows directly from equation (6) which becomes
EC-Dispersion GC-Capillary-Columns
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Compressibility-Effects-GC
;
or
Thus,
(27)
Thus, the
complete HETP equation for a packed GC column, that takes into account the
compressibility of the carrier gas, will be
(28
Dispersion Compressibility-Effects-GC
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Compressibility-Effects-GC
It is seen
that using the average velocity data, the
extracted value for the multi-path term is negative, which is physically impossible (for a capillary
column should be zero or very close to zero). In contrast, the values obtained
from data involving the exit velocity give small positive, but realistic
values for the multi-path term.
In all aspects
of column evaluation and column design in GC, the compressibility of the mobile
phase must be taken into account or serious errors will be incurred.
Figure 10.
De-Convolution of the HETP Curve Obtained Using the Average Mobile Phase
Velocity
Dispersion Compressibility-Effects-GC