Column Length
There are three methods of measuring column length. The first, and the most common, is to measure the distance between the point of connection of the column to the mobile phase supply and the connection of the column to the detector. The second, a more precise method is to measure the actual length of the column packing or, in the case of a capillary column, the actual length of the coated portion of the column. For thermodynamic measurements, the most accurate method is to measure the distance between the point of injection and the end of the column packing or coating. In gas chromatography, capillary columns can be 100m or more in length and, when operated at the optimum mobile phase velocity, can provide very high column efficiencies and consequent high resolution. Conversely, short capillary columns, a few meters in length, operated at very high mobile phase velocities, can produce relatively simple separations extremely rapidly. In liquid chromatography, due to the flow impedance of a packed bed increasing as the particle diameter is reduced, short columns (ca 3 cm long) are packed with the smallest particles (ca 3 m in diameter) and operated at very high mobile phase velocities to provide fast analyses. Conversely, columns packed with particles 20 m in diameter can be used to pack long columns (1m in length) and operated at the optimum mobile phase velocity to produce very high efficiencies and high resolution
Author: RPW Scott
Book:The Thermodynamics of Chromatography
Section:Thermodynamics Other-Methods Chiral-Separations Column-Length
Thus, more
theoretical plates will be required to resolve the enantiomers and thus a
longer column will be necessary. At a temperature of 5˚C and at an ethanol
concentration of 5%v/v, the column need only be about 5 mm long(a length of
column that is impractical to pack and operate). Contemporary columns, shorter
that 2 cm are extremely difficult to operate efficiently.
Figure 24
The Minimum Column Length that will Produce the Required Efficiency
The minimum
column length that will provide the minimum analysis time for this particular
separation is not in the practical range of column lengths normally available.
Consequently, the optimum column length must be a compromise between, that
which is theoretically desirable, and that which is practically possible, and
thus the shortest column available would be chosen.
 
Thermodynamics Other-Methods Chiral-Separations Column-Length
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Alternative-Axes
the column length divided by
the number of theoretical plates in the column) has, for obvious reasons,
become termed the Height Equivalent to the Theoretical
Plate (HETP) and has been given the symbol (H). However, it is seen that
(H) is numerically equal to, , which is, in fact, the variance per
unit length of the column. Thus, the function, , is the variance that
the Rate Theory will provide an explicit equation to define and can be
experimentally calculated for any column from its length and column efficiency.
It follows that the equations that give a value for, (H), the variance per unit
length of the column, have been termed HETP equations
Dispersion Alternative-Axes
Author: RPW Scott
Book:Extra Column Dispersion
Section:EC-Dispersion Small-Bore-Columns
matter. Sometimes it is inevitable
that the detector sensor must be a significant distance from the column exit
and, thus, a relatively long length of connecting tube will be necessary. Under
these conditions, low dispersion tubing, such as serpentine tubing, may be the
solution. Dispersion in the detector sensor is the second largest source of
extra-column dispersion and can only be reduced to a satisfactory level by
reducing the volume of the sensor cell to 2 ml
or less. Reducing the cell length will reduce its response but maintain the
same or similar noise level, thus, reducing the sensitivity (minimum detectable
concentration). Maintaining the same length, but reducing the cell radius, will
maintain the same response, but increase the noise level, also reducing the
overall sensitivity. It is clear a compromise is necessary and this compromise
will depend on the type of detector, UV absorption, fluorescence, etc., that is
employed. Nevertheless, to realize the maximum column
EC-Dispersion Small-Bore-Columns
Author: RPW Scott
Book:Preparative Chromatography
Section:Preparative Packing-Preparative-Columns LC-Columns
. The safe maximum pressure for
any tube can usually be obtained from the tube or column supplier. A slurry is
made of the packing (110% of that needed to fill the column) and placed in the
packing reservoir. The reservoir is rapidly connected to the pump (which must
have both an adequate delivery rate and an adequate operating pressure - the
delivery rate will depend on the column diameter and the applied pressure on
the wall thickness and the column length). The pump is started and the column
exit valve opened and the flow continued until it has been reduced to a
constant value. The flow is then arrested and some packing will remain in the
reservoir which ensures that the top of the column is tightly packed. The
column is disconnected, the packing secured with a suitable frit, and then
connected to the chromatograph.
The major
difference in equipment used for larger scale chromatography lies in the
technology required to pack and maintain high-performance LC columns of
Preparative Packing-Preparative-Columns LC-Columns
Author: RPW Scott
Book:Preparative Chromatography
Section:Preparative Packing-Preparative-Columns LC-Columns
may need to be
determined by experiment. During the initial pressure adjustment some of the
packing passes into the column and forms a lightly packed bed at the bottom of
the column. The exit valve is hen rapidly opened and the sudden flow of gas
packs and compacts the bed at the same time. After packing, the reservoir is
carefully removed so as not to loosen the top of the packing and connected to
the sampling system.
LC Columns
If particle
sizes in excess of 20 mm are used, then
the column can often be dry packed, with appropriate tapping, or, even better,
with longitudinal and radial sonic vibration. The variance per unit length
obtainable from a preparative LC column should be less than 2 particle
diameters (determined using analytical scale samples). It is worth remembering
that (as already discussed) when designing preparative columns, it is better to
obtain the necessary efficiency using a longer column packed with larger
particles, than the converse. The long column
Preparative Packing-Preparative-Columns LC-Columns
Author: RPW Scott
Book:Dispersion in Chromatography Columns
Section:Dispersion Van-Deemter-Equation
efficiency attainable at a given pressure. This is
because, as the particle diameter is increased the column permeability is also
increased allowing a longer column to be used. The permeability increases as
the square of the particle diameter but the variance per unit length only
increases linearly with the particle diameter. Thus, doubling the particle
diameter will allow a column four times the length to be used but the number of
plates per unit length will be halved. Consequently, the column
efficiency will be increased by a factor of two. It is also seen that the
higher efficiencies will be obtained with mobile phases of low viscosity and
for solutes of low diffusivity. Solvent viscosity and solute diffusivity tend
to be inversely proportional to each other and so the sensitivity of the
maximum obtainable efficiency to either solvent viscosity or solute diffusivity
will generally not be large. The approximate length of a column that will
provide the maximum column efficiency
Dispersion Van-Deemter-Equation