A
chromatograph
is an instrument that is used to carry out chromatographic separations. The chromatograph consists of five different entities. The mobile phase supply system, the injection or sampling system, the column and column oven, the detector and the data acquisition and data processing system. In gas chromatography the mobile phase supply system consists of gas tanks, reducing valves, flow controllers and pressure gauges. In liquid chromatography the mobile phase supply system consists of a set of solvent reservoirs, sparged with helium gas to remove dissolved air, a solvent selector valve, a solvent programmer and a high pressure pump. In gas chromatography, the sampling system, in its simplest form, consists of a septum injector and a sampling syringe. In its more sophisticated form, it will include an automatic sampling device that may also be under the chromatograph computer control. Liquid chromatography sampling systems consist of a high pressure sample valve that may also have the sample supplied to the valve from an automatic sampling device. Both gas and liquid chromatography have ovens that are usually temperature programmable, the temperature of the gas chromatographic oven range from 5oC to 350oC and the liquid chromatography oven from about 5oC to 120oC. Gas chromatography usually employs capillary columns whereas packed columns are the most common type of liquid chromatography column. The most common gas chromatography detectors are the flame ionization detector (FID) the Nitrogen Phosphorous Detector (NPD) and the thermal conductivity detector or the hot wire detector (HWD). The three most common liquid chromatography detectors are the UV detector, the fluorescence detector and the refractive index detector (the RI detector). The data acquisition and processing system is very similar for both types of chromatography. They consist of a scaling amplifier and an A/D converter, the output of which is fed to a suitable computer and associated printer. Retention times, retention ratios, peak heights and peak widths are calculated and retention data and quantitative analyses are calculated and then printed out.
Author: RPW Scott
Book:Principles and Practice of Chromatography
Section:Principles Basic-Chromatograph Mobile-Phase
The Basic Column Chromatograph
A
chromatograph consists of five basic units and these units, although possibly
designed differently for different systems, are essential for all types of
chromatograph, including both gas and liquid chromatographs. The layout of all
the five essential units is shown in figure 24. They consists of a mobile phase
supply, a sampling system, a column and column oven including a temperature
controller and temperature programming system. The temperature controller and
Principles Basic-Chromatograph Mobile-Phase
Author: RPW Scott
Book:Liquid Chromatography
Section:HPLC Basic-HPLC
The Basic Liquid Chromatograph
The basic
liquid chromatograph consists of six basic units. The mobile phase supply
system, the pump and programmer, the sample valve, the column, the detector and
finally a means of presenting and processing the results. A block diagram of
the basic liquid chromatograph is shown in figure 1
Figure 1.
The Basic Liquid Chromatograph
HPLC Basic-HPLC
Author: RPW Scott
Book:Gas Chromatography
Section:YES Modern-GC
majority of the added devices and techniques were suggested
or describe in the first three international symposia on gas chromatography
held in 1956, 1958 and 1960. These symposia, initially organized by the
'British Institute of Petroleum' have been held every two years ever since 1956
and the meetings have remained the major stimulus for developing the technique
and extending its capabilities. However, the majority of the techniques and
devices that have been incorporated in the modern chromatograph, were
described, reported, or discussed in the first triad of symposia.
The layout of
the modern gas chromatograph is shown as a block diagram in figure 1.
 
YES Modern-GC
Author: RPW Scott
Book:Capillary Chromatography
Section:Capillary Apparatus Gas-Supply
Apparatus for use with Capillary Columns
The apparatus used specifically with capillary columns differs very little from the general gas chromatograph. Capillary columns require particular sampling systems and also have a more restricted range of suitable detectors than those usable with packed columns, but other than that, the capillary gas chromatograph is identical to that used with packed columns. A diagram of the lay out of a gas chromatograph used specifically with capillary columns is shown in figure 5.
Gas Supply
The gas supply incudes the usual gas tank, reducing valve and flow controller; the pressures
Capillary Apparatus Gas-Supply
Author: RPW Scott
Book:Gas Chromatography
Section:YES Introduction
Introduction
Chromatography,
in one of its several forms, is the most commonly used procedure in
contemporary chemical analysis and the first configuration of chromatography
equipment to be produced in a single composite unit and made commercially
available was the gas chromatograph. Gas chromatography was invented by A. J.
P. Martin who, with R. L. M. Synge, suggested its possibility in a paper on
liquid chromatography published in 1941 (1). Martin and Synge recommended that
the liquid mobile phase used in liquid chromatography could be replaced by a
suitable gas. The basis for this recommendation was that, due to much higher
diffusivities of solutes in gases compared with liquids, the equilibrium
processes involved in a chromatographic process (see
YES Introduction
Author: RPW Scott
Book:Gas Chromatography
Section:YES Gas-Supplies Pure-Air-Generators.
which is also perceived by some as a serious fire hazard despite the fact that
they have been used in laboratories, quite safely for nearly a century.
Supplies from Gas Tanks
Gasses are
stored in large cylindrical tanks fitted with reducing valves that are set to
supply the gas to the instrument at the recommended pressure defined by the
manufacturers. The cylinders are often situated outside and away from the
chromatograph for safety purposes and the gasses are passed to the
chromatograph through copper or stainless steel conduits at relatively low
pressure. The main disadvantage of gas tanks is their size and weight which
makes them difficult to move and replace.
Pure Air Generators.
Air generators
require an air supply from air tanks or directly from the laboratory compressed
air supply. The Packard Zero Air Generator passes the gas through a 0.5 m filter to remove oil and water and finally
over a catalyst to remove hydrocarbons. The
YES Gas-Supplies Pure-Air-Generators.